bunnypartz said:
Hi, I posted earlier that I was getting a couple streaks on the scan,
vertical streaks ... I now believe it's a calibration problem, or perhaps
it is, I have reason to believe it from what I've read so far.
How does a scanner calibrate itself? I mean, where is the calibration
target, and what part is looking at what part ... I just have no idea. If u
know of a site that explains it, I'd be eternally grateful, or somewhat
grateful .. or if u wanna take the time to explain it. There's also
something called an image guide, which ... is related to calibration??
Anywayz I'll continute to search for info on the net in the meantime.
In addition to Wayne's comment, the calibration should have at least two
sections, not just the white response, but a black response as well.
This is usually achieved by scanning a non-reflective or
non-transmissive area.
The output viewing the black level *should* be zero, but it rarely is
due to dark current which flows in the chip and varies with operating
temperature (hence the need to calibrate before the scan. Any output
produced by a CCD cell when viewing the black reference is subtracted
from the output produced when scanning the image to compensate for this
dark current.
The output when viewing the white level *should* be uniform, but not
necessarily peak white. The only variations that should be present are
response and illumination differences across the CCD, and these are
typically quite stable with operating temperature. So the ratio of the
output from each CCD cell to the average of all cells when viewing the
white reference is then multiplied with the output produced when
scanning the image.
In combination, these two operations are just a simple linear correction
for each CCD cell output. In other words, Y = m X + c, Y being the
desired response, X the actual response, m the white correction term and
c the black correction term.
So for the CCD cell number i, the correction applied during the scan is
just Pixel(i) = White(Ave) x (CCD(i) - Black(i)) / White(i)
where:
Black(i) is the output produced by CCD cell i viewing the black surface
White(i) is the output produced by CCD cell i viewing the white surface
White(ave) is the average output produced by all of the CCD cells when
viewing the white surface
CCD(i) is the output of CCD cell i when viewing a particular part of the
image
Pixel(i) is the expected output of an ideal CCD cell i when viewing that
part of the image.
The process of obtaining these black and white reference responses for
all of the cells in the CCD is the calibration process.